We are a flexible packaging manufacturer and make labels for products. We are trying to get SQF certified for the first time this year, so I need a little help in the calibration topic. We use pH meters to test the pH of the inks that we use to print on our film and labels. We have pH testing kits that are provided by the same company that sells the meters to us. Would this be considered internal calibration or would this suffice as an approved method of calibration which would exempt us from outsourcing it?

 

TIA.



Slab

    Grade - SIFSQN

  • IFSQN Senior
  • 438 posts
  • 208 thanks
105
Excellent

  • United States
    United States
  • Gender:Male
  • Location:The Heel of the Boot
  • Interests:Reading (history, science fiction), Photography, drawing,food safety, metrology, TQM, hoplology, etc.

Posted 28 September 2020 - 03:51 PM

This is an internal activity. You may submit the device to an accredited 3rd party for services, but I don't see the reason for this at least according to section 11.2.11 (if I'm fully understanding your question).

 

I would think that for your verifications as long as the buffer is traceable to an international/national known reference standard  than you should be ok. Your vendor should be more than willing to provide all the necessary paperwork for the calibration buffer.

 

 

 

 


Food Safety News  Marine Stewardship Council

 

"Some people freak out when they see small vertebra in their pasta" ~ Chef John


Thanked by 2 Members:

olenazh

    Grade - FIFSQN

  • IFSQN Fellow
  • 1,363 posts
  • 439 thanks
432
Excellent

  • Canada
    Canada
  • Gender:Female
  • Location:Toronto
  • Interests:My job, church, reading, gym, horror movies

Posted 28 September 2020 - 03:51 PM

We use pH meters too: we buy calibrated ones and do internal calibration (against buffer solutions, pH 4 and 7) twice a day. Regarding external calibration (which should be done annually as a pH measurement is our CCP) - we don't do it as our pH meters don't last long, usually 4-6 months, so we buy new ones which're sold calibrated.



Thanked by 1 Member:

SQFconsultant

    SQFconsultant

  • IFSQN Fellow
  • 4,632 posts
  • 1135 thanks
1,126
Excellent

  • United States
    United States
  • Gender:Male
  • Interests:Just when I thought I was out - They pulled me back in!!!

Posted 28 September 2020 - 04:18 PM

You don't have to outsource, you can do so internally - just document it.


All the Best,

 

All Rights Reserved,

Without Prejudice,

Glenn Oster.

Glenn Oster Consulting, LLC -

SQF System Development | Internal Auditor Training | eConsultant

Martha's Vineyard Island, MA - Restored Republic

http://www.GCEMVI.XYZ

http://www.GlennOster.com

 


Thanked by 1 Member:

moskito

    Grade - SIFSQN

  • IFSQN Senior
  • 412 posts
  • 85 thanks
21
Excellent

  • Germany
    Germany
  • Gender:Male

Posted 02 October 2020 - 03:40 PM

Hi,

 

internal, documented work - every day prior first use.

If several people are using the equipment you can use a label (with red or green dot) to demonstrate the the work is already done. You can used this label although to show that the equipment does not work properly.

Same procedure we are using for balances (check weight 1x/d).

 

Rgds

moskito



Thanked by 2 Members:

Charles.C

    Grade - FIFSQN

  • IFSQN Moderator
  • 20,542 posts
  • 5662 thanks
1,544
Excellent

  • Earth
    Earth
  • Gender:Male
  • Interests:SF
    TV
    Movies

Posted 03 October 2020 - 02:28 AM

SQF Clause ?

Is the pH related to Safety ?


Kind Regards,

 

Charles.C


QC/WC

    Grade - AIFSQN

  • IFSQN Associate
  • 48 posts
  • 0 thanks
2
Neutral

  • United States
    United States

Posted 05 October 2020 - 03:15 PM

SQF Clause ?

Is the pH related to Safety ?

 

Clause 13.2.9 Calibration

 

The pH testers test the pH level for our inks that we use to print on our flexographic printing packaging. It is more for quality I guess, rather than safety.


Edited by QC/WC, 05 October 2020 - 03:22 PM.


QA_Spices

    Grade - Active

  • IFSQN Active
  • 15 posts
  • 2 thanks
1
Neutral

  • United States
    United States

Posted 05 October 2020 - 07:41 PM

Hello,

 

We use a pH meter to verify the dilution pH on our detergent for cleaning and sanitizing procedures. As long as you calibrate the device internally with your 4,7 &10 buffers and making sure they are traceable your are covered. Make sure you document it!

 

Hope it helps!