Jump to content

  • Quick Navigation
Photo

Do I calibrate the probe or the thermometer?

Share this

  • You cannot start a new topic
  • Please log in to reply
10 replies to this topic
- - - - -

elaine1980

    Grade - Active

  • IFSQN Associate
  • 23 posts
  • 1 thanks
2
Neutral

  • United Kingdom
    United Kingdom
  • Gender:Female

Posted 18 April 2013 - 05:13 AM

So, my probes are detachable from the thermometer base units. At purchase I alway get a base thermometer with matching probe and a UKAS 5 point calibration certificate. But when they are due for annual external calibration, what do I calibrate? The probe or the thermometer? It's always the base unit that is labelled and gets the unique reference number, but surely the probe attachment is equally as important? Seems a bit crazy to have to calibrate both parts, I'd probably be better just buying new each year for what it would cost me in calibration!



Charles.C

    Grade - FIFSQN

  • IFSQN Moderator
  • 20,542 posts
  • 5665 thanks
1,545
Excellent

  • Earth
    Earth
  • Gender:Male
  • Interests:SF
    TV
    Movies

Posted 18 April 2013 - 06:33 AM

So, my probes are detachable from the thermometer base units. At purchase I alway get a base thermometer with matching probe and a UKAS 5 point calibration certificate. But when they are due for annual external calibration, what do I calibrate? The probe or the thermometer? It's always the base unit that is labelled and gets the unique reference number, but surely the probe attachment is equally as important? Seems a bit crazy to have to calibrate both parts, I'd probably be better just buying new each year for what it would cost me in calibration!


Dear elaine,

Maybe it depends on what you are doing with yr thermometer (thermocouple?) but it's usually economic to purchase one master unit, pay to keep it calibrated at the points you need and self-calibrate all the other units against the master. This is a perfectly acceptable procedure although it can be a bit operationally tricky if you are including temperatures like -18degC or +180degC.

Rgds / Charles.C

Kind Regards,

 

Charles.C


elaine1980

    Grade - Active

  • IFSQN Associate
  • 23 posts
  • 1 thanks
2
Neutral

  • United Kingdom
    United Kingdom
  • Gender:Female

Posted 18 April 2013 - 07:12 AM

Oh dear, blonde moment, of course that is indeed what I'm doing! Problem solved!



Tony-C

    Grade - FIFSQN

  • IFSQN Fellow
  • 4,224 posts
  • 1292 thanks
611
Excellent

  • United Kingdom
    United Kingdom
  • Gender:Male
  • Location:World
  • Interests:My main interests are sports particularly football, pool, scuba diving, skiing and ten pin bowling.

Posted 24 April 2013 - 07:06 PM

Dear elaine,

Maybe it depends on what you are doing with yr thermometer (thermocouple?) but it's usually economic to purchase one master unit, pay to keep it calibrated at the points you need and self-calibrate all the other units against the master. This is a perfectly acceptable procedure although it can be a bit operationally tricky if you are including temperatures like -18degC or +180degC.

Rgds / Charles.C


Hi Charles

I assume that you mean that the probe and thermometer are the one 'master unit'?

Regards,

Tony


Charles.C

    Grade - FIFSQN

  • IFSQN Moderator
  • 20,542 posts
  • 5665 thanks
1,545
Excellent

  • Earth
    Earth
  • Gender:Male
  • Interests:SF
    TV
    Movies

Posted 25 April 2013 - 12:03 AM

Hi Charles

I assume that you mean that the probe and thermometer are the one 'master unit'?

Regards,

Tony


Hi Tony,

Yes. Good point.
IMEX, at time of calibration, a matched/dated “calibration sticker” is attached to each component.

Rgds / Charles

Kind Regards,

 

Charles.C


Rudra

    Grade - MIFSQN

  • IFSQN Member
  • 133 posts
  • 4 thanks
4
Neutral

  • Mauritius
    Mauritius
  • Gender:Male
  • Location:Triolet
  • Interests:Music, reading, surfing, cinema, beach, exercises

Posted 23 May 2013 - 09:45 AM

Hi guys,

 

I use digital  probe thermometers to measure temperatures of dry feeds ( temp ranging between 30 - 95 degrees celcius).

These thermometers are calibrated once a year. My auditor advised me to do a regular in - house verification using ice and boiling water. Does someone have a procedure to do this? How should I carry out this verification.

 

Thanking you in advance,

Rudra



Shyguy77

    Grade - MIFSQN

  • IFSQN Member
  • 120 posts
  • 62 thanks
12
Good

  • United States
    United States
  • Gender:Male

Posted 23 May 2013 - 10:08 AM

 

Hi guys,
 
I use digital  probe thermometers to measure temperatures of dry feeds ( temp ranging between 30 - 95 degrees celcius).
These thermometers are calibrated once a year. My auditor advised me to do a regular in - house verification using ice and boiling water. Does someone have a procedure to do this? How should I carry out this verification.
 
Thanking you in advance,
Rudra


Here you are, hope this is what your looking for and helps you out.

Attached Files


Edited by jpredmore, 23 May 2013 - 10:10 AM.


Shyguy77

    Grade - MIFSQN

  • IFSQN Member
  • 120 posts
  • 62 thanks
12
Good

  • United States
    United States
  • Gender:Male

Posted 23 May 2013 - 10:14 AM

Here is the other attachment. For some reason i couldnt get both of them to post at the same time.

Attached Files


Edited by jpredmore, 23 May 2013 - 10:15 AM.


Thanked by 1 Member:
MCIAN

Rudra

    Grade - MIFSQN

  • IFSQN Member
  • 133 posts
  • 4 thanks
4
Neutral

  • Mauritius
    Mauritius
  • Gender:Male
  • Location:Triolet
  • Interests:Music, reading, surfing, cinema, beach, exercises

Posted 18 July 2013 - 10:37 AM

Hi,

 

There has been a lot of confusion since long between calibration and verification of equipment used to monitor CCPs , for instance, probe thermometer. I have still not yet understood the major difference between calibration and verification...



Charles.C

    Grade - FIFSQN

  • IFSQN Moderator
  • 20,542 posts
  • 5665 thanks
1,545
Excellent

  • Earth
    Earth
  • Gender:Male
  • Interests:SF
    TV
    Movies

Posted 19 July 2013 - 12:26 AM

Hi,

 

There has been a lot of confusion since long between calibration and verification of equipment used to monitor CCPs , for instance, probe thermometer. I have still not yet understood the major difference between calibration and verification...

Dear Rudra,

 

(Thread getting a bit  :off_topic: )

 

Yes it is (and probably always will be) confusing. Should also not forget Validation. :smile:

 

NIST viewpoint on verification –

Attached File  NIST verification.pdf   4.57MB   118 downloads

These other explanations may also be helpful –

 

http://www.instron.u...oogle.co.th/url

http://www.instron.u...rification.aspx

http://www.instron.u...alibration.aspx

 

http://www.ifsqn.com...n-of-equipment/

 

Just to illustrate that you are not the only one, here is a really Geeky solution. Understandable approx. up to end of Pg3 for me. :smile:

http://www.stanford....ESS-Trucano.pdf

 

Rgds / Charles.C


Kind Regards,

 

Charles.C


chiechai

    Grade - Active

  • IFSQN Active
  • 7 posts
  • 0 thanks
0
Neutral

  • Philippines
    Philippines

Posted 07 June 2014 - 12:47 AM

of course its different words with different meanings- calibration and verification :silly:





Share this

0 user(s) are reading this topic

0 members, 0 guests, 0 anonymous users