Validation of MAP gas analyser

Share this

  • You cannot start a new topic
  • Please log in to reply
4 replies to this topic
- - - - -

Archanakolasani

    Grade - AIFSQN

  • IFSQN Associate
  • 25 posts
  • 0 thanks
0
Neutral

  • Australia
    Australia
  • Gender:Male

Posted 30 June 2014 - 04:28 AM

Hi,

 

We bought a new MAP gas analyzer (O2 and CO2 analyser) and wasn't quite sure how to validate the equipment.

 

Can someone please help me how to validate the MAP analyser?

 

Thank you,



jel

    Grade - MIFSQN

  • IFSQN Member
  • 113 posts
  • 33 thanks
13
Good

  • Mexico
    Mexico
  • Gender:Male

Posted 30 June 2014 - 05:23 PM

The equipment and methodology used should be validated as indicated iso 17025: identity, selectivity/specificity, limit of detection, limit of quantification, recovery, working and linera range, recovery, accuracy/trueness, repeatibility precision, intermediate precisión (reproducibility)



Thanked by 1 Member:
Archanakolasani

Snookie

    Grade - FIFSQN

  • IFSQN Fellow
  • 1,625 posts
  • 267 thanks
174
Excellent

  • United States
    United States
  • Gender:Female

Posted 30 June 2014 - 05:39 PM

Your equipment should come with certificate of calibration.  For daily checks we used to check O2 by getting a reading on room air.  It can vary but think it is generally around 20-21%.  Since we also had Nitrogen in the building we would put the tip in a stream of nitrogen to verify that it was not reading O2.  Readings should be stable.  CO2 can be checked by breathing near the tip which should cause a fluctuation.  If you need to validate the equipment you will most likely need to send it to someone such as the manufacturer to verify it is working correctly. 


Posted Image
Live Long & Prosper

Thanked by 1 Member:
Archanakolasani

Archanakolasani

    Grade - AIFSQN

  • IFSQN Associate
  • 25 posts
  • 0 thanks
0
Neutral

  • Australia
    Australia
  • Gender:Male

Posted 30 June 2014 - 11:39 PM

Hi Snookie..Yes our equipment came with a calibration certificate. Doesn't this mean that this is already validated for checking gases and ready to go. With SQF, is the calibration certificate enough for its validation purpose to ensure that the equipment can be used for monitoring or checking the gas mixture in the pack. 

 

Verification is that – Every 12 months – we are calibrating as per manual



jel

    Grade - MIFSQN

  • IFSQN Member
  • 113 posts
  • 33 thanks
13
Good

  • Mexico
    Mexico
  • Gender:Male

Posted 01 July 2014 - 03:00 PM

Validation and calibration are two different concepts, but both are used, and are essential, to provide certainty in the measurements: calibration is to compare the response of the instrument against a standard and establish the chain of traceability; validation, on the other hand, is to ensure that the measurement system is fitness for the intended use. You can have a perfectly well-calibrated system, but it may not be fitness for the intended use



Thanked by 1 Member:
Archanakolasani