Calibration Requirement of Milk UHT plants

Started by , Feb 10 2011 12:11 PM
3 Replies
Dear All,

I request all of you to share the information about the practices followed by your team for the calibration of milk UHT plants. Please cover the below points:-
1.Calibration checks followed during installation periods and certificates of calibration or proof given by the supplier after installation.
2. Validity of Calibration after installation.
3. Calibration requirements of temperature & pressure sensors - What interval for external & internal - and which method is followed for internal calibrations.
4.Also the name of your UHT plant provider (manufacturer details)

If you can also elaborate some more details about the Good maintenance practices of UHT plants or some useful literatures, will be highly thankful.

Regards,
anish
Share this Topic
Topics you might be interested in
Calibration Certificate Expiration Metal Detector Validation vs Calibration (11.2.10.2) Calibrated equipment found out of calibration Infrared Thermometer Calibration Acceptable tolerance of calibration measurement equipment
[Ad]
Dear Anish

In case of any machineries, the manual specifies the requirement of calibration.
Mostly 1 year or below is specified for all machineries
It shall be calibrated by the manufacturer or by the approved vendor.
Pressure gauges, thermometers and temperature gauges requires annual calibration by an external body.
But it can be changed upon the organization requirement. I have seen external calibration schedule of once in 3 month in few dairy plants.
Internal verification is must and it can be verified by an incternal calibrated devices.

In general, an external calibration is required once in a year and internal verification is required once in a month ( depends on devices)

I will try to get more informations about your other requirements...


Regards
Jomy Abraham
.




Dear All,

I request all of you to share the information about the practices followed by your team for the calibration of milk UHT plants. Please cover the below points:-
1.Calibration checks followed during installation periods and certificates of calibration or proof given by the supplier after installation.
2. Validity of Calibration after installation.
3. Calibration requirements of temperature & pressure sensors - What interval for external & internal - and which method is followed for internal calibrations.
4.Also the name of your UHT plant provider (manufacturer details)

If you can also elaborate some more details about the Good maintenance practices of UHT plants or some useful literatures, will be highly thankful.

Regards,
anish

You probably have already dealt with it, but don't forget to include contact time/throughput in your calibration programme

Dear All,

I request all of you to share the information about the practices followed by your team for the calibration of milk UHT plants. Please cover the below points:-
1.Calibration checks followed during installation periods and certificates of calibration or proof given by the supplier after installation.
2. Validity of Calibration after installation.
3. Calibration requirements of temperature & pressure sensors - What interval for external & internal - and which method is followed for internal calibrations.
4.Also the name of your UHT plant provider (manufacturer details)

If you can also elaborate some more details about the Good maintenance practices of UHT plants or some useful literatures, will be highly thankful.

Regards,
anish


Hi Anish

This will vary from plant to plant but generally you will want to calibrate:

Thermometers measuring holding temperature & checking steam barriers
Pressure gauges on steam and sterile air barriers
Flow meters for product flow through your holding tube (since it is usually fixed).

Calibration should be conducted before commissioning and usually verified every 3 months. As long as the methodology is traceable to international standards it can be conducted internally although I like to have an independent check periodically as well.

Regards,

Tony

Similar Discussion Topics
Calibration Certificate Expiration Metal Detector Validation vs Calibration (11.2.10.2) Calibrated equipment found out of calibration Infrared Thermometer Calibration Acceptable tolerance of calibration measurement equipment Area Thermometer Calibration Do calibration weights need to be traceable or ISO certified? Frequency of External Calibration of Analytical Balance and Incubator External Calibration Is it necessary to check the calibration on your mixer and oven timers?