Validation of MAP gas analyser

Started by , Jun 30 2014 04:28 AM
4 Replies

Hi,

 

We bought a new MAP gas analyzer (O2 and CO2 analyser) and wasn't quite sure how to validate the equipment.

 

Can someone please help me how to validate the MAP analyser?

 

Thank you,

Share this Topic
Topics you might be interested in
Process validation of CCP's - What should the frequency be? Shelf life validation for ambient stable sauces Is Scientific Literature Enough for ISO 22000 Validation, or Is Practical Testing Required? Supervising vs Unsupervised Allergen Cleaning Validation Do You Need an Allergen Validation Plan if You Only Produce Nut-Based Foods?
[Ad]

The equipment and methodology used should be validated as indicated iso 17025: identity, selectivity/specificity, limit of detection, limit of quantification, recovery, working and linera range, recovery, accuracy/trueness, repeatibility precision, intermediate precisión (reproducibility)

1 Thank

Your equipment should come with certificate of calibration.  For daily checks we used to check O2 by getting a reading on room air.  It can vary but think it is generally around 20-21%.  Since we also had Nitrogen in the building we would put the tip in a stream of nitrogen to verify that it was not reading O2.  Readings should be stable.  CO2 can be checked by breathing near the tip which should cause a fluctuation.  If you need to validate the equipment you will most likely need to send it to someone such as the manufacturer to verify it is working correctly. 

1 Thank

Hi Snookie..Yes our equipment came with a calibration certificate. Doesn't this mean that this is already validated for checking gases and ready to go. With SQF, is the calibration certificate enough for its validation purpose to ensure that the equipment can be used for monitoring or checking the gas mixture in the pack. 

 

Verification is that – Every 12 months – we are calibrating as per manual

Validation and calibration are two different concepts, but both are used, and are essential, to provide certainty in the measurements: calibration is to compare the response of the instrument against a standard and establish the chain of traceability; validation, on the other hand, is to ensure that the measurement system is fitness for the intended use. You can have a perfectly well-calibrated system, but it may not be fitness for the intended use

1 Thank

Similar Discussion Topics
Process validation of CCP's - What should the frequency be? Shelf life validation for ambient stable sauces Is Scientific Literature Enough for ISO 22000 Validation, or Is Practical Testing Required? Supervising vs Unsupervised Allergen Cleaning Validation Do You Need an Allergen Validation Plan if You Only Produce Nut-Based Foods? Switching Between Tree Nuts: Is New Allergen Validation Needed? Allergen risk assessment and validation for allergen transfer to uniforms Shelf life validation for the fermented batter In-House Gluten Testing – Do COA Results Require Validation? Cleaning validation procedure