Calibration frequency of reference thermometer

Started by , Oct 13 2014 07:28 AM
3 Replies

Dear All,

Does anyone have a reference which state that annual calibration of reference measuring device (eg. thermometer, pressure gauge) is a mandatory?

Thank you for sharing

Avila

Share this Topic
Topics you might be interested in
Calibration requirements for Metal Detectors Metal Derector Calibration Certification Metal Detector Validation vs Calibration How many passes of the wands do we need to do the daily metal detector calibration? Filters, Screens and Strainer Calibration
[Ad]

Dear All,

Does anyone have a reference which state that annual calibration of reference measuring device (eg. thermometer, pressure gauge) is a mandatory?

Thank you for sharing

Avila

 

Dear avila muncar,

 

The calibration interval logically depends on the device, eg via risk assessment. Various examples exist on this forum.

 

However, I daresay you can find specific, published, recommendations for the maximum calibration interval of certain devices, eg by manufacturers.

 

Rgds / Charles.C

 

PS - for a "thermometer" this page may generically answer yr query -

 

http://eziz.org/vacc...er-calibration/

 

attachments - the(1,2)  are probably more scientifically accurate but less directly helpful, the3 is 50/50, the4 is nearer 100%

 

the1 - FDA,1999 - thermometer use-calibration.pdf   17.02KB   145 downloads

the2 - thermometer calibration guide.pdf   450.39KB   146 downloads

the3 - How to decide frequency of calibration for an instruments.pdf   967.65KB   178 downloads

the4 - Support program 4-thermometer use equipment maintenance.pdf   65.99KB   116 downloads

 

 

 

1 Thank

Hello, actually it shall be every 6 months for sensitive parameters of the retort, see "Principles of Canning", "Calibration for Thermal Processing", Campdem Guide CCFRA,

best regards,

Leila

1 Thank

Dear avila muncar,

 

The calibration interval logically depends on the device, eg via risk assessment. Various examples exist on this forum.

 

However, I daresay you can find specific, published, recommendations for the maximum calibration interval of certain devices, eg by manufacturers.

 

Rgds / Charles.C

 

PS - for a "thermometer" this page may generically answer yr query -

 

http://eziz.org/vacc...er-calibration/

 

attachments - the(1,2)  are probably more scientifically accurate but less directly helpful, the3 is 50/50, the4 is nearer 100%

 

the1 - FDA,1999 - thermometer use-calibration.pdf

the2 - thermometer calibration guide.pdf

the3 - How to decide frequency of calibration for an instruments.pdf

the4 - Support program 4-thermometer use equipment maintenance.pdf

thanks for valuable Guide. 

1 Thank

Similar Discussion Topics
Calibration requirements for Metal Detectors Metal Derector Calibration Certification Metal Detector Validation vs Calibration How many passes of the wands do we need to do the daily metal detector calibration? Filters, Screens and Strainer Calibration Thermometer Calibration Calibration CCP Validation ORP Calibration Reference document for setting oven calibration frequency Calibration Certificate Expiration