Test Pieces for Metal Detectors

Started by , Mar 16 2017 09:55 AM
3 Replies

Hi,

Just wanted to ask the required size of metal pieces based on the regulation to be used in testing the efficiency of metal detectors.

Your help is of great help. Thank you

Share this Topic
Topics you might be interested in
Allergen Verification: Do You Need to Test for All Allergens After Changeover? Auditor Dispute: Sampling Technique Competency vs. Test Procedure Effectiveness My sample pH test vs. Lab pH Test -- always differs, why? How Often Should You Test Soap and Sanitizer in Sink Compartments? Looking for a new Listeria Rapid Test for Environmental Monitoring
[Ad]

Depending on the country you are in, you would have to look at their regulations on legal limits for metal in product. We use 2.0 mm SS, 2.0 mm NFe and 1.5mm Fe. Those are a lot smaller than what is legally allowed and my inspector asked me why we didn't get bigger pieces. My reasoning was that we produce flour, and I want to be absolutely sure nothing bigger than that is getting through. 

 

Basically your MD should be able to detect the size you allow or is legally allowed. 

1 Thank

The critical for metal test piece is 1mm for Fe, 1.5mm for NonFe and 2mm for Sus

1 Thank

Hi,

Just wanted to ask the required size of metal pieces based on the regulation to be used in testing the efficiency of metal detectors.

Your help is of great help. Thank you

 

Hi yaba,

 

The MD you use should have a specification (for a given set up) regarding sensitivity, typically defining the smallest (Fe/non-Fe/SS) sphere  "detectable" when passed thru the magnetic field at the worst signal location (typically the central axis). So this sphere dimension can be used to test the "efficiency".

 

The general practical efficiency will particularly relate to factors such as food matrix, eg wet/dry, MD aperture height and sample shape/orientation if non-spherical, eg needle-like/vertical/horizontal.

 

Hopefully, the MD size detection limit (LOD) for yr food matrix will be less than the Regulation.

1 Thank

Similar Discussion Topics
Allergen Verification: Do You Need to Test for All Allergens After Changeover? Auditor Dispute: Sampling Technique Competency vs. Test Procedure Effectiveness My sample pH test vs. Lab pH Test -- always differs, why? How Often Should You Test Soap and Sanitizer in Sink Compartments? Looking for a new Listeria Rapid Test for Environmental Monitoring How should I interpret coliform test results from a 250mL water sample, considering a 1mL transfer to a Hygiena MicroSnap test? Looking for ATP Test Kit Suppliers in Australia Do we need to do a pathogen test each year to confirm that as a dry facility there are no pathogens? how to reply customer's lab test report Commercial Sterility Test