Hi everybody.

I have a question how often do you do calibration for volumetric glassware and for disposable serological pipettes in microbiological labs?

And if you can tell on which document you call on when you make decisions about it?

Thanks in advance

Janie

Share this Topic
Topics you might be interested in
Calibration Certificate Expiration Metal Detector Validation vs Calibration (11.2.10.2) Calibrated equipment found out of calibration Infrared Thermometer Calibration Acceptable tolerance of calibration measurement equipment
[Ad]

Hi everybody.

I have a question how often do you do calibration for volumetric glassware and for disposable serological pipettes in microbiological labs?

And if you can tell on which document you call on when you make decisions about it?

Thanks in advance

Janie

 

Hi Janie,

 

Rarely which is probably incorrect in Principle but the accuracy of most (all?) bacterial counts is surely <<< than any volumetric errors.

 

Certification/accreditation to iso17025 if desired might force a change though.

Hi Janie,

 

Rarely which is probably incorrect in Principle but the accuracy of most (all?) bacterial counts is surely <<< than any volumetric errors.

 

Certification/accreditation to iso17025 if desired might force a change though.

Well, the last year s auditor for iso 17025 insisted that we should do calibration every month. The standard and the appendix that we got with it which is called additional instructions for accreditation of microbiological labs states, that a calibration should be done, but that the freaquncy depends of the needs of each lab. And http://www.european-...ation/ea-4-10-g states that calibration should be done annualy, like my colleagues do in the instrumental part of the lab.

Basicly, I am trying to find any document or any type of defence which I can use do elaborate the fact that I am not doing it monthly, because it is a really way to much unnecessary paperwork.

Well, the last year s auditor for iso 17025 insisted that we should do calibration every month. The standard and the appendix that we got with it which is called additional instructions for accreditation of microbiological labs states, that a calibration should be done, but that the freaquncy depends of the needs of each lab. And http://www.european-...ation/ea-4-10-g states that calibration should be done annualy, like my colleagues do in the instrumental part of the lab.

Basicly, I am trying to find any document or any type of defence which I can use do elaborate the fact that I am not doing it monthly, because it is a really way to much unnecessary paperwork.

Yeah, that sounds awful.

 

I would say glassware should be calibrated immediately after purchase, then at maximum annually, anything more seems ridiculous.

 

Automated pipettors or micropipettes probably more often, but I do mine quarterly and have only two.

Well, the last year s auditor for iso 17025 insisted that we should do calibration every month. The standard and the appendix that we got with it which is called additional instructions for accreditation of microbiological labs states, that a calibration should be done, but that the freaquncy depends of the needs of each lab. And http://www.european-...ation/ea-4-10-g states that calibration should be done annualy, like my colleagues do in the instrumental part of the lab.

Basicly, I am trying to find any document or any type of defence which I can use do elaborate the fact that I am not doing it monthly, because it is a really way to much unnecessary paperwork.

 

Hi Janie,

 

I assume you have some satisfactory calibration data using annual checking.

 

from BRC7 Calibration, 6.4.2 (and likely other standards also) -

 

at a predetermined frequency, based on risk assessment

to a defined method traceable to a recognised national or international standard where possible.

 

I think the above is a satisfactory interpretation of "needs" unless specifically not in compliance with text of yr iso 17025 requirements.

 

If you have the data and would like a short, BRC-oriented, isqn thread link for a generic-type, retrospective, risk-based procedure to validate the annual frequency please inform.


Similar Discussion Topics
Calibration Certificate Expiration Metal Detector Validation vs Calibration (11.2.10.2) Calibrated equipment found out of calibration Infrared Thermometer Calibration Acceptable tolerance of calibration measurement equipment Area Thermometer Calibration Do calibration weights need to be traceable or ISO certified? Frequency of External Calibration of Analytical Balance and Incubator External Calibration Is it necessary to check the calibration on your mixer and oven timers?