2 point vs 3 point thermometer calibration
I have heard of two point calibration may be somehow good enough, say 0& 100. Any suggestion?
Calibration of 18C need to do externally and cause hesitation...
If I need to measure temperature of frozen, chill and cooking temperature, does it mean I need to calibrate 3 point of -18C, 0C & 100C?
I have heard of two point calibration may be somehow good enough, say 0& 100. Any suggestion?
Calibration of 18C need to do externally and cause hesitation...
I think the record low temp for multiplication of bacteria (generally) is -15o c (Planococcus sp)- for pathogens I believe it's closer to -1.5o c (listeria sp) -though i have heard anecdotally as low as -5 (citation needed)
& I personally have never encountered 3 point calib. happening outside of a research lab so....(based on that) from a safety perspective- I wouldn't worry too much about it!
Mike
Dear Mike/Jackson Tsang,
(shortened version sincethe forum system just deleted my first post)(Grrr).
I routinely use 5 point calibration, -18, 0, 10, 90, 100 (degC).
Method - set up one certificated master thermometer and generate sub-masters.
It logically also relates to the accuracy required for the intended application. For example, IMEX most TCs have insuffiicient accuracy for CCPs / Validation if using 0/100 degC only and at intermediate/external values.
(-18degC "area" may be required to monitor freezing capabilities / validate cold rooms)
Rgds / Charles.C
I utilize both 2 and 3 point calibration checks, based on the intended device application (as Charles.C mentioned). If using a thermometer/thermocouple for a narrow temperature range (processing 'room' temp, cook temp, freezing temp) I have typically used a 2 point calibration that brackets the anticipated min/max temp range. When using a device measuring a wider range - cold and hot temps, I use a three point calibration: bracketing the extremes and then one point in or around the middle of that range. I have found this methodology supportable, ensures sufficient accuracy, and not excessively time-consuming.
Keith
better not use the same thermometer for a wide range
choose one for cold and one for hot ranges
if you are using a an accredited thermometer as reference for calibration
this is attached with an accreditation report demonstrating the temperature range of accuracy
I suggest that you prefer more practical reference temperatures, related to common practice e.g. -18 for deep freezer, 4 for refrigeration, 15 for cold cabinets, 63 for hot cabinets, 75 for pasteurization or cooking etc.
in that way your calibrated thermometer will be more related to your application