Could we calibrate a digital thermometer like this way?
For some kinds of digital thermometer (with probe to measure internal temperature and infrared to measure surface one) we could check if they are accurate (such as checking with melting ice) but we can not calibrate them because there is no adjusting part on them. Could we indirecly calibrate like this way?
For example: if temperature measured by probe is 1.5 oF higher than that of melting ice (32oF). A note "substract 1.5 oF" is attached on the thermometer and close to the probe for easy remembering.
When this thermometr is used to measure a temperature of a product, the valule of the temperautre to be written down on the monitoring form is the displayed value minor 1.5 oF (such as displayed temperature is 34oF, the temperature to be written down on the monitoring form is 32.5 oF).
For the infrared, the same way is used.
I thought of sending back to the manufacturer but they advised me to buy the new one because the calibration cost is higher than that of buying the new one. I have asked some regulatory food inspectors and they do not like this idea.
Please advise if you have any experience on this issue.
Thank you
I used to calibrate digital thermometer against master MIG thermometer. Since there is no adjustment, I give correction factor to the reading. In my experience the best digital thermometers are in Japan. They were as good as master thermometer. I donot have the brand name.
Regards
Sujit
Hi All,
For some kinds of digital thermometer (with probe to measure internal temperature and infrared to measure surface one) we could check if they are accurate (such as checking with melting ice) but we can not calibrate them because there is no adjusting part on them. Could we indirecly calibrate like this way?
For example: if temperature measured by probe is 1.5 oF higher than that of melting ice (32oF). A note "substract 1.5 oF" is attached on the thermometer and close to the probe for easy remembering.
When this thermometr is used to measure a temperature of a product, the valule of the temperautre to be written down on the monitoring form is the displayed value minor 1.5 oF (such as displayed temperature is 34oF, the temperature to be written down on the monitoring form is 32.5 oF).
For the infrared, the same way is used.
I thought of sending back to the manufacturer but they advised me to buy the new one because the calibration cost is higher than that of buying the new one. I have asked some regulatory food inspectors and they do not like this idea.
Please advise if you have any experience on this issue.
Thank you
i have seen people using the same procedure with HACCP certification for their digital equipments calibration. but this needs to be calibrated against the standard thermometer. the difference in reading you find can be mentioned and On Job Training can be carried out for monitoring staff and record to be kept for auditor review.
regards.
Riz
I have asked some regulatory food inspectors and they do not like this idea.
The problem is that inspectors expect to see a "official" standardised method despite the fact that the result compared to short-cut methods is probably not much different operationally speaking. They are aware that official methods involve items like traceable master thermometers
Another limitation in ice method only is that it is sometimes necessary to measure temperatures substantially away from 0degC. Inexpensive digital units are often unfortunately not that accurate if you look at their usage specifications.
Rgds / Charles.C
Actually, when I check my digital thermometer in melting ice, I do use a master thermometer. The reason why regulatory inspector does not like is the reading value is not the value recorded in the monitoring form when it is used to check temperature of a product as I gave an example.
Is there any different solution for that issue?
Regards,
Quang
s there any different solution for that issue?
It's called educating the inspector.
I wrote a SOP to explain the correction, ie presented the formula.
Then I validated it.
Then I referenced it on the monitoring form.
Rgds / Charles.C
That means you can't take off a constant nor add a constant by using this method. (I would also argue that you can never make a correction by measuring a single point anyway.) So I completely disagree that it's about educating the auditor!
I would argue that you could go to a different manufacturer and purchase some test caps to be able to do internal calibrations, say once a week and then get them externally calibrated once a year vs. a reference probe. I took this approach and whole kit probably cost around £180 and then there's approx a £40 cost a year. If you're using it for CCP monitoring, I would say it shows lack of management commitment not to spend that kind of money!
I think you misunderstood or, more likely, I was insufficiently clear.
Procedure is to use master thermometer, do external calibration which gives a (hopefully) constant correction for use if no reset master to zero. Then use this master to measure temp. of ice water at same time as insert thermometer "X".
I agree with you, the ice is never zero hence you get a (variable) correction for this parameter. The external calibration people I use claim they use specially crushed ice from distilled water but frankly I don't believe them. I hv also seen other external companies use a purely electronic simulated system, ie black box, which is pre-calibrated.
In the end, I am still not sure it makes that much practical difference to assuming ice is zero degC.
IMEX, the 100degC point is more tricky, heat losses down lines etc.
Rgds / Charles.C
Why do you not believe the calibration company? If you are that unsure why are you using them?!!!