No.
Equipment/ Device Name
Location
Function/ Usage
Impact if Out of Calibration
Likelihood of Drift/ Failure
Risk Rating (H/M/L)
Current Calibration Frequency
Proposed Calibration Frequency
Justification/ Notes
1
Weight Scale
In process, Packing, Quality lab
Ensures correct net weight of packaged product.
Medium – Underweight may cause legal issue; overweight affects cost.
Low
Low
Once week at internal and Once in a year at external.
Keep current
Scale is new, low drift history; legal requirement covered.
2
Magnet Trap
Green coffee & Roaster area
Remove metal impurities from green coffee.
Medium - Loss of Metal Control, Contaminated Products & Equipment Damage.
Medium
Medium
Once in a year.
Keep current
3
Level Transmitter
In process
Used to continuously measure and monitor the level of liquids or bulk solids.
Medium - Overflow/Spillage, Pump Dry Running, Explosion Hazard & Tank Structure Damage.
Medium
Medium
Once in a year.
Keep current
4
Metal Detector
Packing area
Detects metal contamination.
Very high – Food safety risk if malfunctioning.
Low
Medium
Every 4 hours at internal and 6 months at external.
Keep current
Daily performance check effective; annual calibration by external lab.
5
Temperature Gauge
In process, Quality lab
Used to measure and display the temperature of liquids, gases, steam, or object surfaces.
Medium - Explosion Hazard, Equipment Overheating, Personnel Danger, Hazardous Substance Exposure.
Medium
Medium
Once in a year.
Keep current
6
Pressure Gauge
In process, Quality lab
Used to measure and display the internal pressure of gases or liquids within a system.
High - Risk of Explosion or Equipment Rupture, Leakage of Hazardous Materials, Danger to Personnel.
Medium
High
Once in a year.
Keep current
7
Pressure Transmitter
In process
To measure pressure and convert the pressure value into an electronic signal, which is then transmitted to a control system.
High - Explosion or Equipment Rupture, Leakage of Hazardous Materials, Danger to Personnel, Fire and Explosion Hazard.
Medium
High
Once in a year.
Keep current
8
Temperature Sensor/RTD
In process, Quality lab
Measures roasting temperature; critical for product safety and quality.
High – Under/over roasting may cause microbiological risk or affect flavor.
Medium
High
Once in a year.
Keep current
Critical control point (CCP) for process; higher risk due to heat exposure.
9
Moisture Analyzer
Quality lab
Measures product moisture content; critical for shelf life.
High – Too high moisture increases spoilage risk.
Medium
High
Once in a year at external.
6 months
High criticality; found drift in last audit; reduced interval recommended.
No.
Equipment/ Device Name
Location
Function/ Usage
Impact if Out of Calibration
Likelihood of Drift/ Failure
Risk Rating (H/M/L)
Current Calibration Frequency
Proposed Calibration Frequency
Justification/ Notes
1
Weight Scale
In process, Packing, Quality lab
Ensures correct net weight of packaged product.
Medium – Underweight may cause legal issue; overweight affects cost.
Low
Low
Once week at internal and Once in a year at external.
Keep current
Scale is new, low drift history; legal requirement covered.
2
Magnet Trap
Green coffee & Roaster area
Remove metal impurities from green coffee.
Medium - Loss of Metal Control, Contaminated Products & Equipment Damage.
Medium
Medium
Once in a year.
Keep current
3
Level Transmitter
In process
Used to continuously measure and monitor the level of liquids or bulk solids.
Medium - Overflow/Spillage, Pump Dry Running, Explosion Hazard & Tank Structure Damage.
Medium
Medium
Once in a year.
Keep current
4
Metal Detector
Packing area
Detects metal contamination.
Very high – Food safety risk if malfunctioning.
Low
Medium
Every 4 hours at internal and 6 months at external.
Keep current
Daily performance check effective; annual calibration by external lab.
5
Temperature Gauge
In process, Quality lab
Used to measure and display the temperature of liquids, gases, steam, or object surfaces.
Medium - Explosion Hazard, Equipment Overheating, Personnel Danger, Hazardous Substance Exposure.
Medium
Medium
Once in a year.
Keep current
6
Pressure Gauge
In process, Quality lab
Used to measure and display the internal pressure of gases or liquids within a system.
High - Risk of Explosion or Equipment Rupture, Leakage of Hazardous Materials, Danger to Personnel.
Medium
High
Once in a year.
Keep current
7
Pressure Transmitter
In process
To measure pressure and convert the pressure value into an electronic signal, which is then transmitted to a control system.
High - Explosion or Equipment Rupture, Leakage of Hazardous Materials, Danger to Personnel, Fire and Explosion Hazard.
Medium
High
Once in a year.
Keep current
8
Temperature Sensor/RTD
In process, Quality lab
Measures roasting temperature; critical for product safety and quality.
High – Under/over roasting may cause microbiological risk or affect flavor.
Medium
High
Once in a year.
Keep current
Critical control point (CCP) for process; higher risk due to heat exposure.
9
Moisture Analyzer
Quality lab
Measures product moisture content; critical for shelf life.
High – Too high moisture increases spoilage risk.
Medium
High
Once in a year at external.
6 months
High criticality; found drift in last audit; reduced interval recommended.