Verifying the accuracy of an infrared thermometer

Share this

  • You cannot start a new topic
  • Please log in to reply
4 replies to this topic
- - - - -

HARPC

    Grade - AIFSQN

  • IFSQN Associate
  • 40 posts
  • 13 thanks
2
Neutral

  • United States
    United States
  • Gender:Male
  • Location:California

Posted 02 May 2013 - 10:23 PM

Hello All,
Users of infrared (IR) thermometers, how do you verify accuracy?

For our current procedure, we use an NIST certified standard thermometer, beaker, and cold/hot water. Just wondering if anyone else does it differently.


Regards,
Bill



Charles.C

    Grade - FIFSQN

  • IFSQN Moderator
  • 20,542 posts
  • 5665 thanks
1,545
Excellent

  • Earth
    Earth
  • Gender:Male
  • Interests:SF
    TV
    Movies

Posted 03 May 2013 - 07:07 AM

Hello All,
Users of infrared (IR) thermometers, how do you verify accuracy?

For our current procedure, we use an NIST certified standard thermometer, beaker, and cold/hot water. Just wondering if anyone else does it differently.

Regards,
Bill


Dear sci,

I assume "verify" equiv. to 'calibrate".

Not a user but if yr query includes the professionals, I think you will immediately get involved in some high-tech gear and maybe a little advanced Physics. Just try googling "calibration infra-red thermometers" to see what i mean. I have attached the Fluke procedure as an example which you probably know already :smile: .

Attached File  Fluke Infra-red calibration guide.pdf   2.04MB   220 downloads

If you mean the "average" user, there is another nice thread here somewhere which has a quite lengthy input from a seeming merchandiser of such instruments and, from memory, which discusses calibration also. Will link thread later if can search it out.

Rgds / Charles.C

Kind Regards,

 

Charles.C


Thanked by 2 Members:

HARPC

    Grade - AIFSQN

  • IFSQN Associate
  • 40 posts
  • 13 thanks
2
Neutral

  • United States
    United States
  • Gender:Male
  • Location:California

Posted 03 May 2013 - 03:15 PM

Dear sci,

I assume "verify" equiv. to 'calibrate".

Not a user but if yr query includes the professionals, I think you will immediately get involved in some high-tech gear and maybe a little advanced Physics. Just try googling "calibration infra-red thermometers" to see what i mean. I have attached the Fluke procedure as an example which you probably know already :smile: .

Attached File  Fluke Infra-red calibration guide.pdf   2.04MB   220 downloads

If you mean the "average" user, there is another nice thread here somewhere which has a quite lengthy input from a seeming merchandiser of such instruments and, from memory, which discusses calibration also. Will link thread later if can search it out.

Rgds / Charles.C


Hey Charles.C,
Thank you for your response and informative guide on calibration.

But what I was looking for was "Field Verification of Accuracy." The link contains the procedure to do so:

http://www.instrumar...o_umeng0100.pdf

Page 10.

Regards,
Bill


Charles.C

    Grade - FIFSQN

  • IFSQN Moderator
  • 20,542 posts
  • 5665 thanks
1,545
Excellent

  • Earth
    Earth
  • Gender:Male
  • Interests:SF
    TV
    Movies

Posted 03 May 2013 - 10:27 PM

Hey Charles.C,
Thank you for your response and informative guide on calibration.

But what I was looking for was "Field Verification of Accuracy." The link contains the procedure to do so:

http://www.instrumar...o_umeng0100.pdf

Page 10.

Regards,
Bill


Dear sci,

Ok, thks.

Frankly, this basically looks like a “field” calibration to me :smile: . Conceptually similar to comparing a conventional probe with a conventional ref. unit.

Practical differences look especially related to radiation basis of IR / provision of suitable defined temperature source-target. The method described in yr attachment seems a bit more “empirical” as compared to typical procedures for conventional probes.

Quick googling showed alternative field methods of similar basis but intended to improve the source / measuring procedure. One link discussed caveats / criticisms relating to potential errors of usual calibration methods (probably field / non-field). Some of these errors are only too familiar when working with conventional probes.

Examples –

http://www.thermowor...ed-thermometry/
Attached File  gf1 - IR field calibration units - R_02093.pdf   97.49KB   162 downloads
Attached File  gf2 - IR field verification - ITUserGuide.pdf   20.67KB   138 downloads

Sorry if you knew all the above already :smile: . Hopefully some actual users will comment further.

Rgds / Charles.C

Kind Regards,

 

Charles.C


Thanked by 3 Members:

HARPC

    Grade - AIFSQN

  • IFSQN Associate
  • 40 posts
  • 13 thanks
2
Neutral

  • United States
    United States
  • Gender:Male
  • Location:California

Posted 06 May 2013 - 03:28 PM

Dear sci,

Ok, thks.

Frankly, this basically looks like a “field” calibration to me :smile: . Conceptually similar to comparing a conventional probe with a conventional ref. unit.

Practical differences look especially related to radiation basis of IR / provision of suitable defined temperature source-target. The method described in yr attachment seems a bit more “empirical” as compared to typical procedures for conventional probes.

Quick googling showed alternative field methods of similar basis but intended to improve the source / measuring procedure. One link discussed caveats / criticisms relating to potential errors of usual calibration methods (probably field / non-field). Some of these errors are only too familiar when working with conventional probes.

Examples –

http://www.thermowor...ed-thermometry/
Attached File  gf1 - IR field calibration units - R_02093.pdf   97.49KB   162 downloads
Attached File  gf2 - IR field verification - ITUserGuide.pdf   20.67KB   138 downloads

Sorry if you knew all the above already :smile: . Hopefully some actual users will comment further.

Rgds / Charles.C


Hey Charles.C,
Thank you for researching into this matter. The first PDF is a really good guide since it is made from UL. Looks like we will use this method for verifying. Have a great day. Posted Image

Much Appreciated,
Bill