Q1- what does Spec +- 2, 3 or 6 degree C in report means?
Q2- if my acceptance tolerance is only +-1 degree C, how to set the correction? Is it taking max value -2.8 or average all 5 correction readings?
Q1+Q2: In my experience, there hardly is a fixed absolute value as the tolerance of such wide range. Fixing the ratio between tolerance and nominal values makes more sense. The greater a quantity is, the greater its tolerance should follow. The "Spec" column looks like the accepted tolerances. The "Correction" is just the difference between the test and reference reads to compare with "Spec". And "Measurement uncertainty" means the error of the measuring method.
Q4- is about digital thermometer.
I understand internal daily/weekly calibration is using hot water & ice bath.
Method is: insert the master unit and working unti in hot water and ice bath. Then check any different.
For cold temperature calibration, we are inserting it in ice flakes.
I would like to know it should insert in 'ice flakes' or 'ice flakes + water'?
The importance is to keep the isothermal state in your bath. Therefore, you can add ice flakes as long as you stir the mixture well.
Q5- is it acceptable to use calibrated master digital thermometer to calibrate working dial thermometer?
Yes, when you prove the reliability of the master, no matter what its mechanism or its technology is.
Hope this help.