How to determine tolerance

Share this

  • You cannot start a new topic
  • Please log in to reply
4 replies to this topic
- - - - -

pfoude

    Grade - Active

  • IFSQN Active
  • 1 posts
  • 0 thanks
0
Neutral

  • Canada
    Canada

Posted 26 February 2024 - 03:26 PM

How do I determine what my tolerance should be when doing the daily thermometer calibrations before I do a correct action.

 

Example....In boiling water if my thermometer reads 100.6 C.....is this out of tolerance?

At what point do I re calibrate.... 1 degree...0.5 degrees?

What is a common tolerance for a CCP thermometer in food processing?



Scampi

    Fellow

  • IFSQN Fellow
  • 5,514 posts
  • 1515 thanks
1,561
Excellent

  • Canada
    Canada
  • Gender:Not Telling

Posted 26 February 2024 - 03:50 PM

As a general rule of thumb  +/- 1C or 2F degrees, anything more than that it needs to be recalibrated

 

BUT that depends heavily on your process and what you require, sometimes you need to be more accurate


  • G M likes this

Please stop referring to me as Sir/sirs


Thanked by 1 Member:

pfoude

    Grade - Active

  • IFSQN Active
  • 1 posts
  • 0 thanks
0
Neutral

  • Canada
    Canada

Posted 26 February 2024 - 05:39 PM

Is there a rule of thumb for the final product scale...Our pack sizes vary from 5 lb pack to 30 lb's. I would guess the tolerance would be a little higher for larger pack sizes than the smaller ones. Is there a formula I could follow.



jfrey123

    Grade - PIFSQN

  • IFSQN Principal
  • 636 posts
  • 182 thanks
314
Excellent

  • United States
    United States
  • Gender:Male
  • Location:Sparks, NV

Posted 26 February 2024 - 09:03 PM

What is the tolerance on the probe manufacturer's specification?  Last time I wrote an SOP for checking probes, I wrote in the mfg's tolerance for the program and the form we created to check (they're usually around +/- 2F).  Also had to take into account elevation of our facility, because at 4,300 feet, science tells us that water will boil at 204F instead of the 212F we read in textbooks due to pressure at this elevation.  I did a validation to show indeed our water is boiling at 204F, so acceptable test of the probe was 202-206F, and if the probe was found outside of that range, then it was to be replaced.



Scampi

    Fellow

  • IFSQN Fellow
  • 5,514 posts
  • 1515 thanks
1,561
Excellent

  • Canada
    Canada
  • Gender:Not Telling

Posted 27 February 2024 - 04:32 PM

Is there a rule of thumb for the final product scale...Our pack sizes vary from 5 lb pack to 30 lb's. I would guess the tolerance would be a little higher for larger pack sizes than the smaller ones. Is there a formula I could follow.

 Yes, you need to consult weights and measures Canada to determine what the acceptable variance is for each pack size and then ensure your scale(s) is/are capable of that

 

https://ised-isde.ca...ty-declarations

 

https://ised-isde.ca...ty-declarations