Request to help me for any standard in which it is mentioned that Metal detector (CCP) shall be calibrated by external agency and about the frequency for calibration?
Thank
Posted 15 October 2018 - 12:17 PM
Request to help me for any standard in which it is mentioned that Metal detector (CCP) shall be calibrated by external agency and about the frequency for calibration?
Thank
Posted 15 October 2018 - 12:38 PM
I do not know about Indian Law, but in Canada, the UK and the USA, they MUST be calibrated formally once/year
Then you should have some sort of weights so you can check them daily/weekly in house for verification
Please stop referring to me as Sir/sirs
Posted 15 October 2018 - 05:58 PM
Like Scampi said - formal.
Formal, however does not always mean that you must go external.
If your PM person is trained in calibration (and you can prove that) your company should be able to carry this out in-house, however - biggie - check with local authority.
All the Best,
All Rights Reserved,
Without Prejudice,
Glenn Oster.
Glenn Oster Consulting, LLC
Consultants for SQF, ISO-certified payment systems, Non-GMO, BRC, IFS, Lodging, F&B
http://www.GlennOster.com -- 774.563.6161
Posted 15 October 2018 - 06:01 PM
We have to always remember that there is usually another set of laws that governs scales
In Canada, it's Weights and Measures Canada----they require a CERTIFIED person calibrate scales that are used in commerce. So like Glenn said, check the law of the land where your products will be made/sold
Please stop referring to me as Sir/sirs
Posted 15 October 2018 - 07:09 PM
Metal detectors don't have as much of a stringent requirement as scales in terms of requiring state/local certified personnel. Check them daily (or more frequently depending on your process and hazard analysis), prove that you or someone on your staff is trained to calibrate them (by the metal detector company, perhaps?) and you could possibly perform internally. Or just have the manufacturer come on site and complete once yearly or however often is recommended or necessary depending on your machine and your process.
Posted 15 October 2018 - 07:32 PM
We have to always remember that there is usually another set of laws that governs scales
In Canada, it's Weights and Measures Canada----they require a CERTIFIED person calibrate scales that are used in commerce. So like Glenn said, check the law of the land where your products will be made/sold
Sorry Gupta for the confusion---my brain is apprently still on Sunday time..............my statement above does not apply to the metal detector
Please stop referring to me as Sir/sirs
Posted 15 October 2018 - 11:42 PM
However, what is being "calibrated" in metal detector? In some operations manual, the term "caibrate" is also referred to as your equipment is trying to learn the product that is passed thru it. If there are no changes,then the validation or repeated functional check should suffice.
As stated by Glenn, if the personnel are trained they can perform this. But if on occasion the functionality test fails (usually 30 for statistics) or you always change your metal detector settings, then it is better to source out servicing to external party to do some EMI checks.
Posted 16 October 2018 - 03:28 AM
Calibration is a much interpreted term as frequently discussed on this forum,
Seems the FS Standard of OP is unknown but BRC considers that if reference is being made to the usage of test metal spheres, a more appropriate terminology is sensitivity.
Similar comment with regard to "frequency"
Kind Regards,
Charles.C
Posted 02 November 2018 - 01:40 PM
In our last audit, auditor asked me how do you validate (she was looking for external agency/manufacturer or some kind of certificate issued) metal detector.
We do our calibrations/re-calibration thrice a day while packaging. I don't understand how to do validation studies in-house (does anyone have some specific procedures or template ?)
Posted 02 November 2018 - 02:25 PM
In our last audit, auditor asked me how do you validate (she was looking for external agency/manufacturer or some kind of certificate issued) metal detector.
We do our calibrations/re-calibration thrice a day while packaging. I don't understand how to do validation studies in-house (does anyone have some specific procedures or template ?)
Krishna,
As mentioned in the above posts, per law you will need to get it validated once a year by a formal/certified personnel or company. If you have someone certified within your facility/company, you will satisfy the requirement. However, if not, you will need to ask a "Certified" personnel/company and to validate it once a year to satisfy the regulation.
Hope it helps.
Thanked by 1 Member:
|
|
Posted 02 November 2018 - 02:28 PM
I think you've gotten some terminology confused
Validated----this is only done once (preferably before metal detector is even paid for)------will the unit be capable of detecting the nesaccary metals at the required frequency to control the hazard
Verify---daily/weekly QA check (usually QA but not always) to verify detector is working as per SOP
CCP Monitoring ---the operational checks of the metal detector based on the # units passing through it hourly/daily etc
No validation study was performed prior to implementation???
Please stop referring to me as Sir/sirs
Thanked by 1 Member:
|
|
Posted 02 November 2018 - 02:52 PM
I think you've gotten some terminology confused
Validated----this is only done once (preferably before metal detector is even paid for)------will the unit be capable of detecting the nesaccary metals at the required frequency to control the hazard
Verify---daily/weekly QA check (usually QA but not always) to verify detector is working as per SOP
CCP Monitoring ---the operational checks of the metal detector based on the # units passing through it hourly/daily etc
No validation study was performed prior to implementation???
I am not sure when was this installed (I am new in the facility), however most of the equipment are old (without manual).
I on-site quarterly to verify DTEs performing task as per SOP
so I guess I don't have validation study/report/certificate.
Thanks for reply
Posted 02 November 2018 - 03:27 PM
Please stop referring to me as Sir/sirs
Thanked by 1 Member:
|
|
Posted 02 November 2018 - 03:32 PM
Thank you this pdf looks great
Posted 02 November 2018 - 08:32 PM
Hi Scampi,
Thks as always for the offering and definitely no intentions to be over-critical but IMHO this document is sort of gobbledegook.
Seems to have no specific context other than SFCA-related.
IMO this (2014) CFIA document should have included a specific reference to the now classic Codex 2008 document on Validation which it seems to be extracting quotes from. The Codex document has itself sooo much more to say on the matter of "Background".
PS - Indian Gupta = Canadian Krishna ??
Kind Regards,
Charles.C
Thanked by 1 Member:
|
|
Posted 05 November 2018 - 06:15 PM
Hi
CFIA have not ask us to validate metal detector yet ( i mean in last 8 months)
we show that we on-site quarterly and that is sufficient verification for them
However we are planning to go for BRC and some BRC-auditor asked me for annual validation studies/certificate/report for metal detector
Posted 05 November 2018 - 08:20 PM
Hi
CFIA have not ask us to validate metal detector yet ( i mean in last 8 months)
we show that we on-site quarterly and that is sufficient verification for them
However we are planning to go for BRC and some BRC-auditor asked me for annual validation studies/certificate/report for metal detector
Here is a link on this topic that can be helpful: (i would also recommend to see post # 6 for understanding of the requirement)
https://www.ifsqn.co...ctors/?p=105910
IMO: You should inquire the manufacturer of the equipment, as they can advise you how often it needs to be done and if economical and cost effective, they can perform it on site.
Thanked by 1 Member:
|
|
Government Regulatory →
FDA - U.S. Food and Drug Administration →
Should Sanitation Be Considered a CCP? (FDA Audit Experience)Started by OKBaking, 27 Mar 2025 ![]() |
|
![]() |
|
Food Safety Topics →
HACCP - Food Products & Ingredients →
Food preservative as OPRP-CCPStarted by city03, 03 Feb 2025 ![]() |
|
![]() |
|
Food Safety Topics →
HACCP - Food Products & Ingredients →
CCPs for modified atmosphere packaging of fresh meatStarted by Alex A., 09 Jan 2025 ![]() |
|
![]() |
|
Food Safety Topics →
HACCP - Food Products & Ingredients →
CCP for frozen meatStarted by riominicer, 26 Dec 2024 ![]() |
|
![]() |
|
Global Food Safety Standards →
FSSC 22000 Food Manufacturing →
Risk Assessment for Not Using a Metal Detector in Cheese ProductionStarted by LnMl, 28 Nov 2024 ![]() |
|
![]() |
0 members, 1 guests, 0 anonymous users