Posted 09 August 2023 - 02:23 AM

Hi,

 

May I know is there any specified intervals for external calibration (analytical balance & incubator) ? 

 

Can I do it every 2 years because I use it only 2 / 3 times per month and we do internal calibration every month.

 

thanks in advance!

 

  • 0

Tony-C

    Grade - FIFSQN

  • IFSQN Fellow
  • 4,869 posts
  • 1453 thanks
793
Excellent

  • Earth
    Earth
  • Gender:Male
  • Location:World
  • Interests:My main interests are sports particularly football, pool, scuba diving, skiing and ten pin bowling.

Posted 09 August 2023 - 04:11 AM

Hi Fara,

 

Normally external calibration is carried out at a frequency based on a risk.

 

 If the equipment/instrument is carrying out critical checks, such as for food safety, monitoring Critical Control Points or for product legality such as quantity/weight checks then an annual calibration may be required at a minimum.

 

You need to also consider your level of checks/monitoring, historical reliability and any manufacturer’s recommendations.

 

In your case, every 2 years may be acceptable if for example you check the balance with calibrated weights and the incubator temperature with a calibrated NPL thermometer and the calibration certificate you have doesn't have an expiry date.

 

Kind regards,

 

Tony

 


  • 2

IFSQN Implementation Packages, helping sites achieve food safety certification since 2009: 

IFSQN BRC, FSSC 22000, IFS, ISO 22000, SQF (Food, Packaging, Storage & Distribution) Implementation Packages - The Easy Way to Certification

 

Practical Internal Auditor Training for Food Operations - Available via the previous webinar recording. 

Suitable for Internal Auditors as per the requirements of GFSI benchmarked standards including BRCGS and SQF.

 

Practical HACCP Training for Food Safety Teams available via the recording until the next live webinar.

Suitable for food safety (HACCP) team members as per the requirements of GFSI benchmarked standards including BRCGS and SQF.


Thanked by 1 Member:

Fara Dieba

    Grade - AIFSQN

  • IFSQN Associate
  • 26 posts
  • 0 thanks
0
Neutral

  • Earth
    Earth
  • Gender:Female

Posted 09 August 2023 - 06:31 AM

Hi Fara,

 

Normally external calibration is carried out at a frequency based on a risk.

 

 If the equipment/instrument is carrying out critical checks, such as for food safety, monitoring Critical Control Points or for product legality such as quantity/weight checks then an annual calibration may be required at a minimum.

 

You need to also consider your level of checks/monitoring, historical reliability and any manufacturer’s recommendations.

 

In your case, every 2 years may be acceptable if for example you check the balance with calibrated weights and the incubator temperature with a calibrated NPL thermometer and the calibration certificate you have doesn't have an expiry date.

 

Kind regards,

 

Tony

 

 

So it is acceptable if I use mercury thermometer(calibrated by 3rd party) and calibrate the laboratory incubator by myself? 

and I don't need to send my incubator to 3rd party for calibration right?


Edited by Fara Dieba, 09 August 2023 - 06:36 AM.

  • 0

Charles.C

    Grade - FIFSQN

  • IFSQN Moderator
  • 20,542 posts
  • 5704 thanks
1,554
Excellent

  • Earth
    Earth
  • Gender:Male
  • Interests:SF
    TV
    Movies

Posted 09 August 2023 - 10:50 AM

So it is acceptable if I use mercury thermometer(calibrated by 3rd party) and calibrate the laboratory incubator by myself? 

and I don't need to send my incubator to 3rd party for calibration right?

You will presumably also require an appropriate Procedure for calibrating the Incubator (eg temperature variations) and a thermometer externally calibrated at appropriate temperatures.

 

Why use a  mercury thermometer ? ( I daresay cheaper but I would have thought thermocouple more rapid responses and can be miniaturised/automated ).


  • 0

Kind Regards,

 

Charles.C


Thanked by 1 Member:

Fara Dieba

    Grade - AIFSQN

  • IFSQN Associate
  • 26 posts
  • 0 thanks
0
Neutral

  • Earth
    Earth
  • Gender:Female

Posted 11 August 2023 - 09:31 AM

You will presumably also require an appropriate Procedure for calibrating the Incubator (eg temperature variations) and a thermometer externally calibrated at appropriate temperatures.

 

Why use a  mercury thermometer ? ( I daresay cheaper but I would have thought thermocouple more rapid responses and can be miniaturised/automated ).

 

Mercury thermometer act as master when i calibrate the other thermometer. 


  • 0

Charles.C

    Grade - FIFSQN

  • IFSQN Moderator
  • 20,542 posts
  • 5704 thanks
1,554
Excellent

  • Earth
    Earth
  • Gender:Male
  • Interests:SF
    TV
    Movies

Posted 12 August 2023 - 05:48 AM

Mercury thermometer act as master when i calibrate the other thermometer. 

Thanks, that makes sense.

Note that if incubator running at say nominally 36degC, you may require the master/sub-master to have a specific calibration at that value. (The subsequent practical usage  depends on how the external calibrator quantitates/presents their results).


  • 0

Kind Regards,

 

Charles.C