Hello!
I'm at a facility where a MD reject stops the line until the source is found.
I recently visited a company that uses xray. They put rejects aside and revaluate them at the end of the run. Xray false rejects are way more common than MD false rejects, so I guess that makes sense.
Does it make sense to do the same with MD rejects? just put them aside till the end of the run? I know, you're exposing yourself to potentially unpacking pallets.But the metal hits we get are almost always from secondary packaging, and we never find "metal". There's just enough metallic dust somewhere in the cardboard, or they're false rejects. Pretty much the same scenario with the xray.
What are best practice procedures for responding to a MD hit? I'm haven't found any sort of standard guidance.
I know the answer is it depends on your process and risk.
I'd really apprecriate hearing the community's procedures for MD rejects.
Thanks in advance for any insight!
Hello,
I am a technical support person directly responsible for commissioning MD sites and I have been working in this field for over 10 years.
In most cases, this status can be directly identified as the metal detector not being set up well.
In the majority of areas where a metal detector can be used, it can work extremely consistently if the MD is set up correctly.
But a bad reality is that it is very challenging to try to get a metal detector to be very stable.
For a metal detector to achieve a basically satisfactory state, it should have a false detection rate and a missed detection rate of less than 1 in 10,000 respectively.
There are so many things that need to be checked on the metal detector site, the general order of checking is:
1, The inductive signal of a metal detector at the rest state must be low enough!
If the MD's induction signal exceeds the machine's normal signal fluctuations, the cause must be found.
You can't proceed to the next step until you find it.
2, The inductive signal must also be low enough for idle operation. The cause must be found and then the next step.
3, When passing through the product, you must try to minimize the product's own inductive signal.
At the same time, view the inductive signals of the 3 standard test pieces.
Must try to get them to pull as wide a gap as possible, the bigger the better.
Generally speaking, it is best if the inductive signal of the test piece can reach more than 2 times of the product signal.
But that's just in general, it's not a must. This needs to be evaluated on a case-by-case basis. This is actually another separate complex issue that you can ignore for now.
If you can let me know your product, packaging, weight, temperature and your test pieces etc, and if using a LOMA metal detector, I should be able to advise a little more.
Earlier I mentioned that it would be best to go to 1 in 10,000, and in fact our highest record goes to 1 in a million.
In this case, MD will be extremely reliable.
In a typical case, one of our customers complained to us that his metal detector had 9 false alarms in 3 consecutive days.
He asked me to arrive at his facility for further checks as this was unacceptable.
Any questions about metal detectors are welcome!